Below is a sample method timetable for the HP-1100 operated in trap/wash/elute mode.  It has been designed to be used with the associated injection program and valve plumbing.  The trap (column 1) and analytical column (column 2) are in-line while the system injects "protein Drano".  The analytical column is off-line and the trap is in-line while sample is injected/pre-concentrated on the trapping column and salts are diverted to waste.  The contact closure at 4.9 minutes starts the analysis on the MS system.  The analytical column is switched in-line, and sample is eluted from the trap and column with a step to 70% B.  At 13.9 minutes, the analysis is stopped when the contact opens again.  The remaining part of the program re-equilibrates the system back to initial conditions.

Time
Module
Setting
Value
0.00
Binary Pump
flow
0.2 mL/min
0.00
Binary Pump
%B
5%
0.00
Autosampler
External Contact 1
open
0.00
Column compartment
column  switching valve
column 2
1.50
Column compartment
column  switching valve
column 1
4.90
Autosampler
External Contact 1
closed
4.90
Column compartment
column  switching valve
column 2
5.00
Binary Pump
%B
5%
5.10
Binary Pump
%B
70%
13.90
Autosampler
External Contact 1
open
13.90
Binary Pump
flow
0.2 mL/min
14.00
Binary Pump
flow
0.4 mL/min
14.00
Binary Pump
%B
5%
15.90
Binary Pump
flow
0.4 mL/min
16.00
Binary Pump
flow
0.2 mL/min